    Thursday, November 5, 2026 4:45:58 AM
    H2 Detector Always Reads 4% Instead of 1% During Calibration
    7 months ago
    Hoping someone here has run into this before.
    We’re using a GD200-H2 portable H2 gas detector (range 0–4 Vol%), and when we try to calibrate it with a 1% H2 cylinder (balance Nitrogen), the reading doesn’t stay at 1%.
    Instead, it keeps climbing all the way up to 4%, then alarms (red light + beeping).
    So basically, it should read 1%, but it just maxes out.
    Details:
    1. Pump-type portable detector (1–4 Vol%)
    2. Flow rate: < 1.5 LPM
    3. Pressure: < 1 bar (per manual)
    4. Display unit: Vol%
    Also wondering — is it possible to switch from vol% to ppm on this device? Any ideas appreciated.
    3
    7 months ago
    AA wrote:
    Details:
    This usually points to a flow/pressure issue rather than the detector itself. If your flow is anywhere near 1.5 LPM, that might already be too high. A lot of these detectors are calibrated for much lower, controlled flow rates.
    4
    7 months ago
    AA wrote:
    Details:
    Yeah, and I’d also ask how long you're applying the gas?
    If you're not giving it enough time to stabilize, the reading can overshoot like that. Pretty common with pump-type detectors.
    3
    7 months ago
    AA wrote:
    Details:
    One thing that stands out, you mentioned “less than 1 bar.” That might actually be the problem. The detector usually expects a stable pressure, not just “below a limit.” If the pressure is fluctuating, readings can go all over the place.
    7 months ago
    Agree with the others. I’d double check the full setup: regulator, tubing, leaks, etc. In my experience, when a detector pegs at full scale during calibration, it’s often something in the setup rather than the detector failing.
    7 months ago
    Got it. For reference, our flow is under 1.5 LPM and pressure is under 1 bar as suggested in the manual.
    7 months ago
    AA wrote:
    Details:
    Thanks for the details. Based on your situation, please verify the following:
    The pressure must be stable at 1 Bar; it may not fall below 1 Bar.
    The flow rate should be stable at 750 ml/min, not close to or exceeding 1.5 LPM;
    The gas application time must exceed 90 seconds to ensure proper stabilization.
    If all the above conditions are confirmed and the issue persists, please arrange a call with our technician for further assistance.
    Regarding your question about units: switching is only available between ppm and mg/m3. It is not possible to switch between %vol and ppm on this detector.
    7 months ago
    That helps a lot, thanks. We’ll try again with a stable 1 bar supply, lower the flow closer to 750 ml/min, and ensure we apply gas for over 90 seconds.